Regarding Meter/strip accuracy:
Most people rarely use the control solution to check their meters/strips. Here's the way I check and adjust my testing process:
The control solution is a constant. Meters and strips I've found will vary according to temperature; cold - higher readings - warm - lower readings.
On your test strip bottle there will be a control test range. On my current OneTouch bottle that number is 115-153mg/dL. That means that when you use the control solution in a test...anything between 115-153 is a viable number. Therein lies the ±15% of accuracy that the manufacture is given. In other words...you add 115 and 153 together, then divide it by 2 and you get 134. 134 minus 15% equals 114. 134 plus 15% equals 154.
So, this means that 134mg/dL is what that meter and test strip ideally should read if you use the test solution at the right ambient temperature.
Now, let's say I test a new bottle of test strips at the ideal temperature range. That range being the middle of the allowed temperature ranges for the meter and test strips. For my meter it's 43 to 111ºF...but my test strip bottle says 'Store test strips in vial only, below 86ºF'. Anyway...
Let's say I use the control solution to test a new bottle of test strips. The control solution test comes out 141mg/dL. Since the bottle reads 115 - 153 that means that the strips for that bottle read 7 points higher than what they should be...or, 5.5%.
From then on, when I test my whole blood, I figure that test is about 5.5% higher than it should be for that bottle.
As to temperature differences; In January I kept getting higher readings than usual. Couldn't figure out why. Using my usual method of Sherlock Holmes' deduction - "When you eliminate the impossible whatever remains, however improbable, must be the truth." I concluded it must be the temperature.
Upon rising in the morning, room temperature of 55ºF I took a FBG test. 132mg/dL. I waited an hour and took another test. 131mg/dL. The next day...did the same thing. FBG tests remained similar.
The next day I took a FBG test and it read 138mg/dL. I put the meter and the bottle of test strips in my pants pocket for an hour. My hour-later FBG test? 110mg/dL. Same thing the next day. I call that empirical evidence.
FWIW